Hi all,

I'm upgrading my spider vanes to the TS Optics double carbon spider (Teleskop-Express: TS-Optics Carbon Double-Spider for 295-310 mm inside tube diameter (teleskop-express.de)). The aim here is to be a bit stiffer and hopefully hold collimation better plus to tighten my diffraction spikes. I will also us copper foil on the vanes for the secondary heater rather than wires.

I've been trying to find advice on the orientation of the vanes relative to the focuser. Everything I see online points to one of the vanes being inline with the centerre of the focuser. However on my OOUK CT10 the stock spider is offset at 45 degrees to this. That makes collimation more tricky as the adjustments tilt/twist the secondary not relative to the optical axis (the two left thumbscrews are not parallel to the optical axis).

Question am I overthinking this just install in the existing holes or should I redrill the fixings in a more usual pattern relative to the focuser?

Thanks in advance

Matt   

20201125_095333.jpg

spider.JPG

Edited by mbalkham
Link to comment
Share on other sites

In the new one I'm not sure if 45 degree rotation works either because of the offset of the two thumbscrews? Will wait to see when it arrives! I've also reached out to TS Optics for advice.

 

ts-double-spider-1000.jpg

Edited by mbalkham
Link to comment
Share on other sites

  • 2 weeks later...

So testing tonight produced a funny set of diffraction spikes. Does this mean I haven't managed to get it square? or is this some other issue (don't think it is collimation or focus but could be tilt, focuser tube alignment, something else?). Also screenshot of ASTAP CCD Inspector for same image.

 

Canopus_Light_007_d.jpg

Screenshot 2020-12-10 at 23.43.41.png

Edited by mbalkham
Link to comment
Share on other sites

I think the split spikes are caused when the secondary isn't centered in the tube, i.e. opposing pairs of vanes don't form a straight line.

 

Looking at your images, it barely looks out... but maybe something worth checking?

 

 

Capture1.PNG

Capture2.PNG

Edited by CraigT82
Link to comment
Share on other sites

I think you might be right. I remeasured the spacing and it looks like my 4 holes are not equally spaced around the circumference (not by much). I was not far off centre but obviously not close enough and that was pulling vanes offline a little. 

I've filled the holes and will redrill tomorrow above the last lot. Will triple check all the measurements and accuracy of the pilot holes before redrilling.

I think I'm parallel to the primary. I'll check again. Will also check the focuser alignment again. 

Link to comment
Share on other sites

  • 5 months later...
On 11/12/2020 at 09:41, mbalkham said:

I think you might be right. I remeasured the spacing and it looks like my 4 holes are not equally spaced around the circumference (not by much). I was not far off centre but obviously not close enough and that was pulling vanes offline a little. 

I've filled the holes and will redrill tomorrow above the last lot. Will triple check all the measurements and accuracy of the pilot holes before redrilling.

I think I'm parallel to the primary. I'll check again. Will also check the focuser alignment again. 

Did you ever get the spikes tighter / to single spikes?

I'm considering getting one of these for my CT10.

Link to comment
Share on other sites

@Starflyer

I'm getting closer to getting it dialled in. My current iteration uses a stainless steel band (see photos) but I've tried just marking and drilling and couldn't get it accurate enough. Any difference in distance to the primary twists the spider (and the spikes don't align) and any slight difference in spacing around the tube misaligns the vanes with a similar result. I considered trying to spring load the attachment to ensure my bolt tightening was equal and not distorting the spider but adopted adjustment under the stars to try and dial in as best I could.

What I've learnt so far:

  1. The hole alignment is critical and drilling them right is tricky! (See below for my latest thoughts). Not only to they need to be equidistant from each other but also from the primary. Any error means that the vanes don't end up in the right alignment and the diffraction spikes never perfectly align.
  2. Changing the rotation by 45degrees has made collimation slightly easier as I suspected above with alignment of the adjustment screws with the optical arrangement.
  3. My mirror isn't equidistant from the end of my tube) or more likely the end of my tube isn't square) so I can't measure from the end of the tube for anything critical
  4. A wide band of paper/thin card is prob the best way to mark 4 equidistant holes. Take the band off and fold in half twice and mark those lines. The width of the band means it will be parallel to the tube
  5. My current solution was to clamp a (actually 2) stainless steel hose clamps over the top of the marked paper band and mark the 4 equidistant lines and the end of the hose clamps. Then to remove the hose clamps, remeasure/check that the holes are equidistant and punch and drill the 4 spider holes in the centre of the band. When you tighten the clamp it will square itself to the tube was my theory
  6. I aligned one of the holes placing the bolt through the clamp and OTA and then filing out the remaining 3 holes to perfectly align.
  7. I have filled the old holes with thickened epoxy with black dye. They look ok (from a distance!)
  8. Measure the spider to check it is roughly centred. Tighten the for bolts of the spider finger tight. 

I did the collimation after installation as per usual (catseye and glatter combo for me). I did also check that the focuser was square and centred using a hole opposite the focuser drilled with he paper trick. After that I used the laser to check and adjust focuser alignment.

Under the sky I honed my centring of the spider. If it's not dead centre it distorts the vanes and the diffraction spikes aren't right (they diverge away from the star). My Rho image below shows some residual diffraction spike divergence (and was before I installed my steel band). I've improved it a little more since I took that image (see the other test image of Rigel I think).

My process for adjusting under the stars was:

  1. Point at a bright star, focus and take an image
  2. Adjust the spider screws towards the diverging diffraction spike. If your sensor is aligned to the spider this should be easier. If at 45 degrees I suspect you'll need to adjust all four screws simultaneously (and refocus between adjustments)
  3. In my case i could adjust left and right without affecting the focus too much but up and down requires a refocus (as it moves the secondary towards/away from the sensor.
  4. After a bit of fiddling I got to a point I was relatively happy (I do still want to make some final tweaks). This adjustment is likely to have affected your collimation so recollimate at this point.

All in all a lot of hassle! I'm getting there. I recon one more night under the stars and I'll have it pretty dialled in.

Good luck if you do decide to make the mod! I recon sticking with the factory spider holes might be a lot easier. Then again you might highlight some issues you never thought you had when you put the new spider in! Def holds collimation better and I think my diffraction spikes will be better (eventually!).

 

 

20210530_140727.jpg

20210530_140648.jpg

Spider Test.jpg

NvSgCNkewEgO_16536x0_b9muqi8S.jpg

Edited by mbalkham
Link to comment
Share on other sites

On 06/06/2021 at 23:21, mbalkham said:

@Starflyer

I'm getting closer to getting it dialled in. My current iteration uses a stainless steel band (see photos) but I've tried just marking and drilling and couldn't get it accurate enough. Any difference in distance to the primary twists the spider (and the spikes don't align) and any slight difference in spacing around the tube misaligns the vanes with a similar result. I considered trying to spring load the attachment to ensure my bolt tightening was equal and not distorting the spider but adopted adjustment under the stars to try and dial in as best I could.

What I've learnt so far:

  1. The hole alignment is critical and drilling them right is tricky! (See below for my latest thoughts). Not only to they need to be equidistant from each other but also from the primary. Any error means that the vanes don't end up in the right alignment and the diffraction spikes never perfectly align.
  2. Changing the rotation by 45degrees has made collimation slightly easier as I suspected above with alignment of the adjustment screws with the optical arrangement.
  3. My mirror isn't equidistant from the end of my tube) or more likely the end of my tube isn't square) so I can't measure from the end of the tube for anything critical
  4. A wide band of paper/thin card is prob the best way to mark 4 equidistant holes. Take the band off and fold in half twice and mark those lines. The width of the band means it will be parallel to the tube
  5. My current solution was to clamp a (actually 2) stainless steel hose clamps over the top of the marked paper band and mark the 4 equidistant lines and the end of the hose clamps. Then to remove the hose clamps, remeasure/check that the holes are equidistant and punch and drill the 4 spider holes in the centre of the band. When you tighten the clamp it will square itself to the tube was my theory
  6. I aligned one of the holes placing the bolt through the clamp and OTA and then filing out the remaining 3 holes to perfectly align.
  7. I have filled the old holes with thickened epoxy with black dye. They look ok (from a distance!)
  8. Measure the spider to check it is roughly centred. Tighten the for bolts of the spider finger tight. 

I did the collimation after installation as per usual (catseye and glatter combo for me). I did also check that the focuser was square and centred using a hole opposite the focuser drilled with he paper trick. After that I used the laser to check and adjust focuser alignment.

Under the sky I honed my centring of the spider. If it's not dead centre it distorts the vanes and the diffraction spikes aren't right (they diverge away from the star). My Rho image below shows some residual diffraction spike divergence (and was before I installed my steel band). I've improved it a little more since I took that image (see the other test image of Rigel I think).

My process for adjusting under the stars was:

  1. Point at a bright star, focus and take an image
  2. Adjust the spider screws towards the diverging diffraction spike. If your sensor is aligned to the spider this should be easier. If at 45 degrees I suspect you'll need to adjust all four screws simultaneously (and refocus between adjustments)
  3. In my case i could adjust left and right without affecting the focus too much but up and down requires a refocus (as it moves the secondary towards/away from the sensor.
  4. After a bit of fiddling I got to a point I was relatively happy (I do still want to make some final tweaks). This adjustment is likely to have affected your collimation so recollimate at this point.

All in all a lot of hassle! I'm getting there. I recon one more night under the stars and I'll have it pretty dialled in.

Good luck if you do decide to make the mod! I recon sticking with the factory spider holes might be a lot easier. Then again you might highlight some issues you never thought you had when you put the new spider in! Def holds collimation better and I think my diffraction spikes will be better (eventually!).

 

 

20210530_140727.jpg

20210530_140648.jpg

Spider Test.jpg

NvSgCNkewEgO_16536x0_b9muqi8S.jpg

Thanks for the very in-depth reply.  I was planning to use the existing holes, I prefer the spikes at 45° to horizontal anyway.

You seem to have it very nearly dialled in, I'd be happy with that.

I've seen CNC machined spiders and thought about one of these but can't find one to fit the CT10. Zero adjustment, so it should be exactly centered, but if you need to adjust you're out of luck.

Thanks again,

Ian

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue. By using this site, you agree to our Terms of Use.