  1. Paul,

    Understood.

    I assume your markings on the edge of the slit holder should match the necessary orientation of the slit plate. That would remove ambiguity.

    Sorry if I've "muddied the waters" - my only experience is with the Custom designed plates.

    Ken

     

  2. Paul,

    I assume(d) the current Jeulin reflective slit plate was a copy of the Custom design plate that OVIO manufactured for me. When viewed from the telescope side, the numbers appear correct, the chrome coating and slits are on the rear face - towards the collimator.

     

    OVIO custom 001.JPG

    OVIO A3017KA Master Drawing.jpg

  3. Paul,

    I’d like to think all the amateurs who build a LowSpec, and any other spectrograph will put them to regular use. We need as many active spectroscopists as possible.

    There are more challenges and opportunities in spectroscopy than there are observers.

  4. For completeness:

    Coronado use the ITF (or Maier equivalent), Lunt went for a Blue Schott BG 38 filter - unfortunately they both can fail....

    The Maier replacement has been used by MANY solar observers over the past seven years and I'm not aware of any failures!

     

    • Like 2
    • Thanks 1
  5. Louise,

    No, definitely more like the Chinese one....

    I mounted it in a small project box, with 12V socket and on/off switch. This sat on the outside of the housing and powered the Neon lamp fitted in a perspex block which flipped down onto the slit gap.

     

  6. The ITF (Induced transmission Filter) is usually the filter element placed before the final narrowband blocking/ sort filter in an Ha set-up.

    These are subject to failure.....luckily we found a suitable replacement from Maier Optronics.

     

    • Like 1
    • Thanks 1
  7. A spectrograph definitely needs a reference lamp.

    As a minimum a Neon, this provides calibration around Ha which is an important wavelength for astronomers. The RELCO is also good and give a wider coverage.

    When I started with the Spectra-L200 "IKEA" kit, I had the same reference lamp issues.....

    I found at the time that the hand held fluoro "inspection" lamps had a small 12V to 110/220V inverter built in, so I used to raid the hardware shops and buy them up, and strip them for the inverter and the 12V plug.

    The inverter looked very similar to this one:

    https://www.banggood.com/150W-Inverter-Boost-Module-150W-DC12V-Step-Up-Board-Frequency-Square-Wave-p-1278570.html?rmmds=detail-left-hotproducts__3&cur_warehouse=CN

    This initially powered a neon lamp. I used about 100 of them (!!) with no failures, no issues, in fact I'm still using one of the early build inverter/ neon set-ups today, almost ten years later.

    This is shown in "Astronomical Spectroscopy for Amateurs" p160 and p202. There's also a "non commercial"  circuit diagram for a suitable inverter.

     

  8. Louise,

    Yes, every time... thermal  effects, flexing in the housing etc etc etc can impact on the accuracy of measurements. It's not unusual to take a reference image before and after a long series of target exposures.

    Fitting a reference lamp - check the alignment and orientation to the slit gap. Ideally the lamp should be on the optical axis as near the slit gap as possible.  On the Spectra-L200 I used a flip mirror to direct the reference lamp to the slit gap.

     

  9. 28 minutes ago, Thalestris24 said:

    Here is my S10:

    S10_only_slant_flip.png.dbae2babd2b6af460efae0d35fd7c2d1.png

    CFL:

    CFL_only_slant_flip.png.fa8d7559f2b40ca1909429e0c17347cb.png

    Both:

    1618414973_S10CFL_slant_flip.png.569227dcfefaabf06b3094f09c1959f5.png

    The S10 is a bit of a jungle at the red end! Presumably that's mostly neon plus other stuff.

    Anyway, I'll give it another go starting from the cfl tomorrow though different people quote slightly different wavelengths for some of the lines...

    Louise

    Louise,

    Use the Buil reference image I supplied for the CFL - it has been used successfully by many.

    Need to check back in the messages...but I assume you have the line/ wavelength data for the S10?

    ken

     

  10. Louise,

    Softly softly....

    1. Start with the CFL as a "reference image" (place up front as #1 on the list). Us the reference image I posted with the lines identified and mark/ calibrate  all the obvious lines you can see.

    This will give you a pretty good calibration of your instrument set-up. Once you have done this then.....

    2. Import the S10 image (as #2 image on the list). Right click on the image, Properties/ Calibration/ select "use calibration from the first profile" / copy. This will then calibrate the S10 image form the "reference image" of the FCL.

    3. If you now want to positively identify any/ all the lines in the S10 spectrum - use the mouse to move over the lines (Crop the X axis if needed) to match lines with the reference that Eric gave. If you want, you can then lable some of the lines - for future reference - Click on the "label" icon, in the pop down screen enter the type of lable you want to use (check the BASS manula for details) When done, save the chart as a .png (Chart/ save chart to file)

    Re-reading your message..the S10 data is just that...data which you manually have to select and enter, you can't just superimpose the image.....

    .Ken

     

    • Like 1
×
×