Moravian G2-8300 has ceased to function


tomato

Recommended Posts

I took my G2-8300 out of mothballs tonight to do some Ha imaging but although it connects OK to the software (NINA via ASCOM or SIPS with the Moravian native driver) after a couple of good subs I just get a dark screen or one with a light band down the LHS.

Powering down and reconnecting sometimes brings a couple of good subs back but then it reverts back again to the dark subs. The cooler control is fine and it is receiving a steady 12.9v as it has always done. I’ve swapped USB cables, no difference.

The question is, is it a dead parrot? It’s nearly 9 years old so it would have to go back to the Czech Republic with all the hassle that entails and if it isn’t a simple, cheap repair it probably isn’t economically viable.

Anybody else had a KAF 8300 camera exhibit this behaviour?

  • Sad 1
Link to comment
Share on other sites

47 minutes ago, tomato said:

I took my G2-8300 out of mothballs tonight to do some Ha imaging but although it connects OK to the software (NINA via ASCOM or SIPS with the Moravian native driver) after a couple of good subs I just get a dark screen or one with a light band down the LHS.

Powering down and reconnecting sometimes brings a couple of good subs back but then it reverts back again to the dark subs. The cooler control is fine and it is receiving a steady 12.9v as it has always done. I’ve swapped USB cables, no difference.

The question is, is it a dead parrot? It’s nearly 9 years old so it would have to go back to the Czech Republic with all the hassle that entails and if it isn’t a simple, cheap repair it probably isn’t economically viable.

Anybody else had a KAF 8300 camera exhibit this behaviour?

Is the shutter sticking?

  • Thanks 1
  • Confused 1
Link to comment
Share on other sites

@markse68 and @Adam J, you guys were right on the money, when I took the camera off I found this:

7F1139FB-E5AD-4217-82C8-DD25F687192D.thumb.jpeg.6eeed62b8a9e4dc1117e0cbe3e0d304d.jpeg
 

Further inspection revealed that the shutter vane had lost it’s clearance on one side, you can see where it had been catching in this photo:

119227BB-EDFB-4BC0-B790-F110A472EBAF.thumb.jpeg.fe696507c5fb73d68601b17fca1f9452.jpeg
 

I just bent the vane very slightly to restore the clearance and it appears to be operating 100% reliably once more. There are some ribbon cables associated with the integrated filter wheel which I may have misrouted when reassembling after a filter change, so it’s probably all down to me, apologies if I have trashed Moravian Instruments’ quality reputation.

Thanks for your feedback, I’m a happy CCD imager again👍😉

  • Like 6
Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue. By using this site, you agree to our Terms of Use.